  1. #14
    Join Date
    Jul 2000
    Location
    Guayaquil, EC
    Posts
    13,584
    Post Likes
    Quote Originally Posted by jimp View Post
    Oil was in the middle of sight glass when I removed the old compresor, as it was in the new compressor when installed.
    Well that's another possibility eliminated.

  2. #15
    Join Date
    Nov 2011
    Location
    East Side
    Posts
    5,376
    Post Likes
    How did it fail?

    Electrical or mechanical?

    Did you cut the old one open and look at the bearings?

  3. #16
    Join Date
    Feb 2005
    Location
    Central Florida
    Posts
    1,253
    Post Likes
    Thread Starter
    Its definitely mechanical.

    It has the MCS controller, When the chiller starts it warm the water to 70F, at this point it adjusts the cooling by the demand frequency of the magnet. Today when I put the new compressor in, the FSE told me they had offsets in the hot gas bypass controller and the low pressure safety cutout.

    It has two low pressure setting. When the chiller comes on the suction pressure drops to 45 PSI before the liquid line solenoid opens, the low pressure safety cutout is 10 PSI.

    Last edited by jimp; 10-13-2018 at 06:32 PM.

  4. #17
    Join Date
    Feb 2005
    Location
    Central Florida
    Posts
    1,253
    Post Likes
    Thread Starter
    Before I start the new compressor tomorrow the FSE will be there, and he needs to correct the offsets or have new a transducer and thermistor.

    The FSE's that monitor the sites from another state sometimes make ignorant decision to keep the magnet running.

    I know recently that one FSE put a higher number in the HP cutout because it was tripping out, but not this site.

  5. #18
    Join Date
    Dec 2004
    Location
    Chicago area
    Posts
    6,796
    Post Likes
    Whats an FSE ?


    Sent from my iPhone using Tapatalk

  6. #19
    Join Date
    Dec 2014
    Posts
    2,308
    Post Likes
    Quote Originally Posted by heatingman View Post
    Whats an FSE ?


    Sent from my iPhone using Tapatalk
    Field service expert
    Frantic service expert
    Failed service expert
    Factory silly entity
    Fat senile engineer.

    Really not sure

    Sent from my SM-G930V using Tapatalk

  7. Likes heatingman liked this post
  8. #20
    Join Date
    Feb 2005
    Location
    Central Florida
    Posts
    1,253
    Post Likes
    Thread Starter
    Quote Originally Posted by heatingman View Post
    Whats an FSE ?


    Sent from my iPhone using Tapatalk

    The MRI world calls their service techs, field service engineers.

  9. #21
    Join Date
    Aug 2002
    Location
    Southold, NY
    Posts
    19,015
    Post Likes
    Ask for better yet Demand copies of all operating parameters. Your problem will be in the reports.

  10. #22
    Join Date
    May 2006
    Posts
    3,316
    Post Likes
    is that the old compressor or the new one to be installed and why do you have no front plate in Maine we have to have them

  11. #23
    Join Date
    Feb 2005
    Location
    Central Florida
    Posts
    1,253
    Post Likes
    Thread Starter
    So I replaced the failed compressor on 10/15. Had them default all setting in the MCS, they replaced the thermistor for the HGB and defaulted all settings.

    I haven't heard a beep from the site, and they paid to change the second compressor.