    thermo scientific -86c

    im a chiller guy so this is kinda outa my wheel house headed to look at one on monday and noticed while reading the iom that there is a rs232 port is there any software i should have installed on my laptop. also any tips or common occurrences with equipment like this would be greatly appreciated. dont know much about the issue other than not making temp

  2. #2
    Join Date
    Aug 2002
    Location
    Southold, NY
    Posts
    15,832
    Post Likes
    Cascade systems.

    One system removes the heat from the case, the other removes heat from the first.
    There tech support is good, giver them a call.

  3. #3
    Join Date
    Mar 2007
    Location
    In the Conservatory with a Lead Pipe
    Posts
    14,719
    Post Likes
    These things seem to always be in some kind of third basement bunker with no cell service for me. If they have training on these things it would be nice but I've never seen it offered.

  4. #4
    Join Date
    Aug 2002
    Location
    Southold, NY
    Posts
    15,832
    Post Likes
    A lot of On The Job training.

    Thermo Scientific does offer classes but i could never get the boss to pay for 5 days.

  5. #5
    Join Date
    Mar 2005
    Location
    Pearland Texas
    Posts
    387
    Post Likes
    Most of the ones I work on are older versions. There is a probe coming out of the heat exchanger you can connect to your temperature meter. The temperature has to get low enough for the 2cd stage to come on. Ditto on their tech support. They are very helpful.

  6. #6
    Join Date
    May 2014
    Location
    Bay Area California
    Posts
    18,904
    Post Likes
    If it's an electrical problem, then troubleshoot like anything else.

    If it's a refrigerant problem, the high stage is a weigh in charge, and the low stage is charged to a static pressure when everything is at ambient. Easy to overcharge the low stage if things are cold.
    A skilled Tech would solder a relay on that board and call it good to go.

  7. #7
    Join Date
    Mar 2005
    Location
    Pearland Texas
    Posts
    387
    Post Likes
    Sometimes the evaporators on the low side get oil logged. I have to turn them off to let them warm up. When I turn them back on the oil will go back to the compressor.

  8. #8
    Join Date
    Dec 2015
    Location
    Michigan
    Posts
    32
    Post Likes
    What model?
    Has it been moved recently?
    What refrigerants being used?

    Have the customer shut it down over the weekend. This way the refrigeration systems
    can stabilize and come up to ambient room temperature when you arrive.

    If they have gauges on refer systems look and I.d. high stage and the low stage.
    Machine I.d. tag should have refrigerants used and how much. Usually they are:

    High stage: R-404a or R-507 newer stuff. Older stuff R-502, R-12, and R-22
    Low stage: R-23 and R-508b " ". Older stuff R-503, R-13, and R-13b

    80% of the time these systems have leaks. Look for oil and bring leak detector.
    Prepare yourself for the cost of low temp refrigerant. Virgin DuPont R-23 $ 125.00 per pound 10 and 20 pounders
    R-508b $ 150 to $175 per pound and same size containers sold. $ 150.00 deposit on hard cylinder.

    If your system is small a.k.a. bench top with fractional h.p. comps. then system charges are very small
    less than 2 lbs. Good luck!

  9. #9
    Join Date
    Sep 2017
    Location
    north louisiana
    Posts
    31
    Post Likes
    Thread Starter
    Thanks guys I found a locked up high stage compressor will be replacing compressor and adding oil separator as per factory recommendations.