    BlueVac Problem

    just got it this week and man is it great.
    Used it a few times on a couple recovery jugs. With a old vacuum pump that was so called "bad".

    Pulled a 100lb cylinder down to 80 and it only rose to 150 in 10 mins. will pull down just a brass coupler from trutech down below 100 in 20 seconds and 2 mins maybe less to below 20.

    But now it is reading 0 microns at atm pressure.
    Changed the battery and still doing it.
    Any idea why and how to resolve problem?
    thanks

  2. #2
    Join Date
    Sep 2008
    Location
    Western PA
    Posts
    25,946

  3. #3
    Join Date
    Jun 2009
    Location
    South Carolina
    Posts
    501
    ran it on a pump for about a hour and it finally went away and started readling properly.been using it prolly 3 hours total today and it did not happen again.

    Took it a while to reset must have had oil or something on the sensor.

  4. #4
    Join Date
    Sep 2008
    Location
    Western PA
    Posts
    25,946
    Quote Originally Posted by klee7013 View Post
    ran it on a pump for about a hour and it finally went away and started readling properly.been using it prolly 3 hours total today and it did not happen again.

    Took it a while to reset must have had oil or something on the sensor.
    If it has oil on the sensor, it will say OIL on the screen and requires a cleaning with alcohol.

  5. #5
    Join Date
    Jun 2009
    Location
    South Carolina
    Posts
    501
    Thats weird then.
    It was at 0 microns not hooked up to anything for about 10 mins- then hooked it to a vacuum pump and got stuck working on something else cut it back on it and was reading properly.

    I have used it alot today because i was playing with our vacuum pump graveyard and vacuuming out recovery tanks(9). 3 out of 4 pumps that were "bad" (sad to say before we used micron gauges) are good now. changed the oil and tightened a couple of fittings and one capacitor later. They all pull down between 20-60 microns in under a minute. We got one frankesteen pump up and running with parts off of the one that i couldn't get working. It has a bad seal too bad its not just the oil resovior seal its coming from underneath.
    They are all robinair pumps i know one is a 6 cfm pump the other ones the labels are wore off. For doing that i got an okay to order some new micron gauges for our techs looks like they will get some yellow uei / (bluvacs) from jmart and can get rid of some of the old ones that are not very accurate at all.

    Thanks BluVac

  6. #6
    Join Date
    Mar 2007
    Location
    wedged in freezer shelf
    Posts
    7,197
    I never tried but is zero a selection on the alarm setting. Maybe it was stuck where you would have set the alarm
    “If You Can Dodge A Wrench You Can Dodge A Ball”

  7. #7
    Join Date
    Jun 2009
    Location
    South Carolina
    Posts
    501
    Does the LTE have a timer?

  8. #8
    Join Date
    Mar 2007
    Location
    wedged in freezer shelf
    Posts
    7,197
    I don't think so
    there goes that idea huh
    “If You Can Dodge A Wrench You Can Dodge A Ball”

  9. #9
    Join Date
    Aug 2010
    Location
    Florida
    Posts
    251
    klee7013,

    The behavior you described is not normal. Keep an eye on it. If it happens again, call me at my shop and I'll repair it for you. The number is on my website http://accutools.com.

    Apologies for the problem.

  10. #10
    Join Date
    Jun 2009
    Location
    South Carolina
    Posts
    501
    Will do Joey. Thanks for the help

  11. #11
    Join Date
    Feb 2007
    Location
    Fremont, Indiana
    Posts
    1,755
    Mine (lte) keeps getting stuck in calibration mode.
    Either says "soon”, or good, but will not let me exit. I have to pull the battery.
    I have cleaned it, and will try to calibrate if it will let me!

    sent from my Samsung Galaxy Note
    Member of the "Work Exchange Program"
    "Will work for knowledge"

    "Everyone is a genius. But if you judge a fish on its ability to climb a tree, it will live its whole life believing that it is stupid"
    A Einstein

  12. #12
    Join Date
    Aug 2010
    Location
    Florida
    Posts
    251
    Quote Originally Posted by kamersoutdoor View Post
    Mine (lte) keeps getting stuck in calibration mode.
    Either says "soon”, or good, but will not let me exit. I have to pull the battery.
    I have cleaned it, and will try to calibrate if it will let me!
    This sounds like a problem with the keypad. Call me on Monday to arrange repair.

  13. #13
    Join Date
    Nov 2004
    Location
    up in the hizzy
    Posts
    1,440
    I bought my Thermal micron gauge 20 years ago, no problems to the date.

Page 1 of 3 123 LastLast

Posting Permissions

  • You may not post new threads
  • You may not post replies
  • You may not post attachments
  • You may not edit your posts
  •  
Comfortech Show Promo Image

Related Forums

Plumbing Talks | Contractor Magazine
Forums | Electrical Construction & Maintenance (EC&M) Magazine
Comfortech365 Virtual Event