PDA

View Full Version : calibratable thermometer



wannafreeze
01-10-2008, 10:43 PM
Is there a dependible digital thermometer out there that I can calibrate it myself when needed?

wannafreeze
01-18-2008, 08:19 PM
to answer my own question..
I bought the fieldpiece ST4.It's pretty good.You can offset both readings from the front screws.the only minus is operating ambient on the low end is only 32F.

MechAcc
01-20-2008, 09:06 AM
Fluke 52 and 54

rimek
01-20-2008, 09:38 AM
Most of the handhelds are calibratable, but if you're only checking in an ice bath, then you are only doing 1 point. Better to have annuals done by an instrumentation shop or authorized agent and get certifications for 3-4 points, then you have accountability as per ISO. Most of our clients need that cert for the FDA anyway, when we come and do the PM or repair for their equipment.

icehouse
02-02-2008, 06:13 PM
Also your DMM by the power company. Had this done by LIPA, when a customer lost product in "brown-out" causing compressor problems I showed the cert in court. Judge said "Pay the man" !